WalkerNelson678
Registered
Hello All,
At my company, we are developing a low-risk "wearable" medical device. The device is accompanied by a wireless charging pad, which is the subject of my question.
The charging pad is rated for 5V 0.5A DC input, and therefore we are claiming that it is NOT a mains part. However, the most obvious use case is that the operator plugs the pad into a phone charger to supply power.
After discussion (and a previous post here) we have convinced ourselves of, and have high confidence in, the decision that that safety testing applicable to mains connected parts (one example is 60601-1-2 -- section 8.9 -- table 5, although there are others from 60601-1 itself) will not apply to us. Mainly this is for two reasons:
However, we are getting some pushback from consultants / test labs to consider using the IEC 6238-1 power supply as a MOOP for the charger pad. This is acceptable according to 60601-1 clause 8.5.1.3 with the caveat that the power supply only constitute one MOOP (to the best of my knowledge).
Although this action is technically allowable, I worry that by including the power supply as a MOOP, we open the door for an FDA reviewer to question our exemption from the aforementioned "mains part" tests. I am thinking that if we use the PSU as a MOOP, a reviewer could claim that it is part of our equipment, and therefore will requires us to go back and retest for all of the AC power port tests which we originally exempted ourselves from.
So finally, my question:
Are these two concepts/statements mutually exclusive in 60601?
Thanks,
~Walker
At my company, we are developing a low-risk "wearable" medical device. The device is accompanied by a wireless charging pad, which is the subject of my question.
The charging pad is rated for 5V 0.5A DC input, and therefore we are claiming that it is NOT a mains part. However, the most obvious use case is that the operator plugs the pad into a phone charger to supply power.
After discussion (and a previous post here) we have convinced ourselves of, and have high confidence in, the decision that that safety testing applicable to mains connected parts (one example is 60601-1-2 -- section 8.9 -- table 5, although there are others from 60601-1 itself) will not apply to us. Mainly this is for two reasons:
- We do not manufacture the USB power supply
- We state in the user manual that only the use of a power supply conforming with IEC 62368-1 is acceptable
However, we are getting some pushback from consultants / test labs to consider using the IEC 6238-1 power supply as a MOOP for the charger pad. This is acceptable according to 60601-1 clause 8.5.1.3 with the caveat that the power supply only constitute one MOOP (to the best of my knowledge).
Although this action is technically allowable, I worry that by including the power supply as a MOOP, we open the door for an FDA reviewer to question our exemption from the aforementioned "mains part" tests. I am thinking that if we use the PSU as a MOOP, a reviewer could claim that it is part of our equipment, and therefore will requires us to go back and retest for all of the AC power port tests which we originally exempted ourselves from.
So finally, my question:
Are these two concepts/statements mutually exclusive in 60601?
- My device is usable with any 62368-1 PSU which supplies 5V 0.5A DC, therefore, the PSU is not part of my equipment and I should not have to perform tests which are required for a mains part (again, such as 60601-1-2 section 8.9, table 5).
- A 62368-1 compliant PSU forms one of the MOOPs for my device
Thanks,
~Walker